Capintec- Data-5 / CRC-55tPET / Add for desired Radio Isotope
 글쓴이 : 운영자
조회 : 5,671  

Add for desired Radio Isotope


Please note that the Capintec CRC-55T PET Dose Calibrator is not designed for the measurement of standard isotopes.

 

This is due to the pressurization (5ATM) that is used for the ionization chamber. A general purpose calibrator such as the CRC-55TR uses an ionization chamber pressure of 12ATM.

 

(This is why for the CRC-55T PET there are no calibration factors listed for standard isotopes.)

Recommended Cal. Nos. for use with a CRC-55T PET when measuring standard [non- PET] isotopes. The Cal. Nos. are for estimation only.

 

 

--------------------------------------------------------------------------------------------------

1] For the measurement of Tl-201. I-125. I-123 and Ga-67 with the CRC-55T PET

     you will need to obtain samples where the activity is already known or calibrated and then follow the procedure in the attachment.

 


 
   
 

 

Dz